Capt Jim Wyatt
Hanging at the 10 Foot Stop
Staff member
ScubaBoard Business Sponsor
Scuba Instructor
Calibration must be done using a gas with a "known" O2 %. We know what the % of O2 is in atmospheric air, however temperature & humidity can alter the reading by as much as 1%. Also the analyzer has an inherent error of somewhere around 1%, so you could be off as much as 2% by using only atmospheric air without compensating for humidity and temperature.
Wave the analyzer around in atmospheric air to get a reading of somewhere around 21%. Then blow air slowly from what you think is a known source (scuba tank with air in it) into the analyzer, if the difference is less than 1% you can be certain that the tank does indeed have only air in it. AS you blow air into it (Flow rate should be 2 liters/minute) set the calibration knob to 20.9%. You can then have a very high degree of confidence that the instrument is calibrated correctly.
Wave the analyzer around in atmospheric air to get a reading of somewhere around 21%. Then blow air slowly from what you think is a known source (scuba tank with air in it) into the analyzer, if the difference is less than 1% you can be certain that the tank does indeed have only air in it. AS you blow air into it (Flow rate should be 2 liters/minute) set the calibration knob to 20.9%. You can then have a very high degree of confidence that the instrument is calibrated correctly.