I have an OEII and am wondering whether this is normal as I have never experienced this with any other units of the same model.
After equilibrating to ambient temp and then turning on and calibrating to ambient air, the reading will then begin to drift down, reading under 20 within a minute or two - if I then recalibrate to air it will then start to drift down again. When initially turning on, I have seen readings between 19.4 and 21.9.
Calibrating with air and then checking against 100% Oxygen gives a reading in the 95.6 to 96.5 range.
Analyzing the same tank of 32% shows a variance of up to 1% after calibration with ambient air (done some time apart).
Is this normal or within spec? I am only concerned as I have never seen this amount of variation with another unit.
I have an OEII and am wondering whether this is normal as I have never experienced this with any other units of the same model.
After equilibrating to ambient temp and then turning on and calibrating to ambient air, the reading will then begin to drift down, reading under 20 within a minute or two - if I then recalibrate to air it will then start to drift down again. When initially turning on, I have seen readings between 19.4 and 21.9.