A cell checker also highlighted that true linearity is a bit of a myth, or at least is something that is either taken out of context or not actually found in the wild. If I calibrate at 1.0, I get acceptably accurate readings in the range that matters from .8 to 1.2 with errors in the 1% to 2% range (at an actual PPO2 of 1.2, a 2% error will give you a reading of 1.17 - no big deal). However while I may have .05% less error at a PPO2 of 1.0, and around 1%-2% at 1.2, the deviation from the theoretical maximum will always continue to increase as I get further away from that calibration point. It's not uncommon to see a 5% or 6% error at a PPO2 of 1.6, which means at an actual PP02 of 1.6 I'm only going to see a reading of 1.54 to 1.52 on the computer. Until I figured that out, I blamed the low reading at 20 ft on un-swept volume in the unit and/or on less than perfect O2 flushing.